- https://manual.kikusui.co.jp/P/PAS_USERS_E4.pdf
复制代码
• After calibrating the offset and full scale of the voltage and current, be sure to press the SET switch to exit the calibration.
If you switch to a calibration of another item or turn off the POWER switch before pressing the SET switch, that calibration is invalid.
• In the calibration mode, the unit outputs voltage or current used to calibrate the offset or full scale.
The unit outputs approximately 10 % of the rated output during offset calibration and approximately the rated output during full scale calibration.
请务必按下SET开关退出校准。 |
|